Dear Changhee,
you need the condenser front lens Achr. 1,30 (304332:032.24/6) or Achr. 1,3 pol
together with its matching rump optics (see right image and the arrows, green is interchangeable, red not): The rump optics of the condenser head achr. apl. 0,9 does not work together with the higher aperture front lenses! They came with their own rump optics. Sometimes, condenser front heads are offered separetely from their rump by people who are not aware of that. Such purchases are useless. Also, the early JENAVAL condensers had different threads on head and rump compared to what was offered later. The versions from 1982 - 1984 had even female and male threads reversed from the later versions! They have an engraving in white (not the silk- screen print in silver colour).
So You need this: The front and rump optics shown, are the latest versions discernible by the
silver coloured silk- screen print of "Achr. 1,3" or "Achr. 1,3 pol", made until 1990. Depicted here:

Regarding the image quality: In phase contrast, the higher aperture condenser front optics Achr. 1,30 (used with self- made, modified Ph- ring stops) just gives a brighter image (factor ~ 1,8x) not much better resolved -if at all:

Marine Diatom (J. LIEDER; Mr. Heinz Fegert lege artis). JENA 250-CF Apochromat HI 100x/1,40 Phv oo/0,17-A with phase contrast and condenser head Achr. apl. 0,9 (Colour image with LED- light source made by Stephan Hiller, no image manipulation/ processing after acquisition)
In DIC there is a difference, but that would be a very lenghty story to explain here, especially what is all needed to make it work. And still, I am not sure, if it is worth all the effort ... The condenser head with N.A. 0,9 is also more than excellent for almost everything I could think of (except a few diatoms here).

Freshwater nano-flagellate
Bodo angustus. Length is approx. 15 µm. JENA 250-CF Apochromat HI 100x/1,40 oo/0,17-A with DIC and condenser head Achr. 1,3 pol immersed with IMMERSOL 518 F (Colour image with PANASONIC LUMIX GX-9 and PANASONIC TTL DMW-FL 580 used as micro flash and made by Jürgen Stahlschmidt, no image manipulation/ processing after acquisition)
If You cannot access the JENAVAL Achr. 1,3 front and rump optics, a home made adapter for the much easier to get CARL ZEISS, Oberkochen condenser achr. apl. 1,40 H, DIC will also work beautiful with such a condenser on the JENAVAL in brightfield or DIC. As I have no such adapter any longer (its gone since more than 25 years), I cannot depict it for You here ...
Regards
Michael